Ionization Chambers
Introduction
These "ion chambers" are nothing more than a bare wire stuck through a hole
into a metal can! No special gas or sealing is required. For best performance it is
probably a good idea to add a desiccant to the inside of the can to keep the humidity low.
(I didn't!) Build one; its really simple!
When ionizing radiation (ultra-violet light, x-rays, etc.) pass through a gas,
collisions with the gas molecules produces ion pairs, typically charged molecules and free
electrons. If an electric field is present, the ions will move apart, each moving in
opposite directions along the electric field lines until they encounter the conductors
that are producing the electric field.
An ion chamber is an extremely simple device that uses this principle to detect
ionizing radiation. The basic chamber is simply a conducting can, usually metal, with a
wire electrode at the center, well insulated from the chamber walls. The chamber is most
commonly filled with ordinary dry air but other gasses like carbon dioxide or pressurized
air can give greater sensitivity. A DC voltage is applied between the outer can and the
center electrode to create an electric field that sweeps the ions to the oppositely
charged electrodes. Typically, the outer can has most of the potential with respect to
ground so that the circuitry is near ground potential. The center wire is held near zero
volts and the resulting current in the center wire is measured.
The voltage required to sweep the ions apart and to the center wire or outer can before
a significant number of them recombine or stick to a neutral molecule is usually under 100
volts and is often just a few volts. In fact, if the voltage is above a couple of hundred
volts, the speeding electrons will produce additional ion pairs called "secondary
emissions" giving an enhanced response. Geiger tubes operate at even higher voltages
with a special mixture of gasses and exhibit a sudden and very large discharge for each
ionizing particle. But below 100 volts the only current is the ions produced by the
radiation. The resulting current is extremely low in most situations and detecting
individual x-rays is difficult, especially with ordinary air at atmospheric pressure.
Usually the capacitance of the electronics connected to the center wire smoothes the
individual pulses too much for detection even when feedback is used to greatly reduce the
time constant. These room-pressure chambers therefore respond to the average level of
ionizing radiation and do not provide "clicks" like a Geiger counter tube.
Homebrew
Sensitive homemade ion chambers for detecting nuclear radiation are fairly easy to
build but the circuitry is tricky and should only be attempted by "seasoned"
experimenters - the currents are likely to be well below 1 pA unless there is a serious
nuclear war in progress! (The simple version is
"beginner friendly"!) Special electronics is needed at the front end, typically
called an "electrometer" circuit, which produces an output voltage in proportion
to the input current. The electrometer must have a very low bias or leakage current to
avoid masking the desired signal and the intrinsic impedance of the amplifier must be
extremely high. The input impedance of the electrometer may be fairly low, however, using
feedback to convert the tiny current into a usable voltage.
Older designs used special electrometer tubes like the 5886 which requires only 10 mA
at 1.25 volts for the filament and about 10 volts for the plate. These tubes are great for
the experimenter because they are relatively immune to static discharge and they consume
about the same amount of power as a typical transistor stage. Some electrometers use
vibrating capacitors or mechanical choppers to convert the tiny DC currents into AC before
amplification to avoid DC bias and leakage problems. Newer circuits typically use MOSFETs
or Electrometer grade JFETs in the front-end. MOSFET op-amps usually contain protection
diodes which can be responsible for several picoamperes of leakage at room temperature and
a fairly steep increase in leakage as the temperature increases but in some ion chamber
applications this extra leakage is tolerable. Non-protected MOSFET front-ends are easily
damaged by static electricity and special low-leakage protection diodes are usually added.
Low current JFETs like the 2N4220 give respectable performance and the types intended for
electrometer applications like the 2N4117A are quite impressive, exhibiting leakage well
below 1 pA. They have the added benefit of being significantly less sensitive to static
electricity than unprotected MOSFETs. Full ESD precautions must be observed with any of
these approaches!
Big Resistors!
As mentioned earlier, most electrometer circuits use feedback to reduce the effective input impedance and to direct the tiny input current through a very large feedback resistor such that a reasonable voltage is produced at the output. The feedback resistor must be quite large, however. If the input current is 1 pA and the feedback resistor is 100 megohms, the output voltage will only be 100uV. Special resistors measuring in the millions of megohms are available but are usually difficult for the experimenter to obtain (see http://www.ohmite.com/catalog/v_rx1m.html , for example). Recently, I stumbled upon a quantity of 13,000 megohm resistors - send me an email if you want one. That same 1 pA would give 13 millivolts which is easily amplified with ordinary op-amp circuitry.An Experimental Circuit
An extremely sensitive circuit was desired that didn't require special
resistors and that didn't fail every time a slight ESD mistake was made and the result is
the experimental circuit shown below. It uses a 2N4117A as the input amp and another
as the feedback resistor. If one studies the tiny curves supplied in the data books and
uses a little "extrapolation" and imagination, the leakage of the 2N4117A with
the drain and source connected together can be seen to have a slope equivalent to about 75
million megohms! There is unit-to-unit variation and it is necessary for the input JFET to
have lower leakage than the feedback JFET so the circuit is not for everyone. (If the
input JFET leakage is higher, the output voltage will be very low. Simply swap the JFETs.)
The FETs are easily damaged, too, which can lead to frustration when the "best
one" gets zapped. The circuit will not be particularly accurate since the actual
feedback resistance is not known (nor linear) but the experimental ion chamber is not
easily characterized anyway. Despite the circuit's shortcomings, it is extremely sensitive
and surprisingly stable
A "good one" might drift only 0.1 fA in a day if
ambient conditions are relatively constant. (That corresponds to about 10 mV drift on the
output.) The short-term variation is below 1 mV which corresponds to 0.01fA! If the ion
chambers really work, this circuit should be able to see the current!
The input FET (the one on the right) and the two transistors form an
"error" amplifier that attempts to maintain the drain voltage at 10 volts (set
by the resistor divider in the emitter of the NPN). If current flows out of the ion
chamber causing the gate voltage to rise, the drain will begin to drop and the voltage on
the collector of the NPN will go up. This rise will decrease the current in the PNP and
thus lower the output voltage. The voltage drop across the first "resistor" FET
will increase and more current will flow through it - nearly all of the ion chamber
current, in fact. There is sufficient loop gain that the input voltage does not change
very much and most of the ion chamber current flows through the feedback FET. The zener in
the source of the input FET moves the gate voltage operating point up above ground so that
dual polarity supplies are not needed. The output voltage should be a few volts, perhaps 3
or 4, depending upon the relative leakage of the FETs. If the voltage gets too near 6
volts, the sensitivity will drop and the response will become more logarithmic (which
might be useful for some applications). If the voltage is too low, the circuit might
"bottom out" and loose control. There isn't much that can be done to set the
operating point expect swap out FETs! The glass around the FET leads must be VERY clean.
Use a good solvent to remove any contaminants.
Ion Chambers
The first experimental ion chamber was made with a zinc can from a D-cell battery and
an old 8-pin glass-to-metal header as seen in the photo below. The two FETs were mounted
inside the chamber with the theory that this would eliminate the problem of connecting the
extremely high impedance probe to the outside world without creating leakage paths to
ground. The problem with the concept is that the transistor bodies and leads compete with
the wire for the free ions! Carefully painting the transistor bodies and legs with
conformal coating helped but the circuit will not tolerate the coating around the base of
the transistor - it is too conductive! (In retrospect, the transistors should reside in
their own can with the sense wire passing though a hole into the ion chamber which is what
the schematic shows.) The pickup wire should be thin and near the center of the can to
keep the capacitance low so the response time is as short as possible. When power is first
applied, it can take a very long time, maybe 20 minutes, before the circuit settles out to
a steady reading. At first, about 150 volts was applied to the can but it was soon
discovered that only a few volts are adequate and a 9 volt battery was used instead. The
15 volt power supply voltage should be fine for most ion chamber sizes. If the voltage is
too low, the readings will be low as the ions have time to recombine before being swept to
the electrodes.
To test the chamber, a 1.3" diameter disk of radioactive material (a
calibration disc from an old Geiger counter) was leaned against the can. The voltage
changed a few 10s of millivolts but I quickly lost interest in this chamber when the lid
slipped and zapped the FETs. I had already spotted an old mint can at the back of
the workbench which I liked better for a chamber. (See pictures below.) An audio
connector was added to the center of the 3" dia. tin can and the FETs were mounted
directly to the pins. A ring of wire was used for the center electrode. The insides
were washed well with a solvent and then dried with a hot air gun before the base was
added and tack-soldered. (I should have removed the plastic coating on the inside of the
can.)


Little feet were added to the bottom of the can so that I could easily
slip my radiation disk underneath without disturbing the chamber. This ion chamber gave
gratifying result: the little radiation source gave an output voltage change of about 70
mV which was very large compared to the meter wander of about 2 mV.
At this point in the festivities, I decided to try a crude calibration.
Really crude. My calibration reference was a Heathkit Geiger counter which has a meter
that reads counts per minute and mR/ hour. The scales are a little suspicious since the
CPM scale is an exact power of 10 bigger than the mR/hr scale. (0.3 mR/hr = 300 CPM
on the X1 scale, for example.) It is entirely possible that the Geiger tube dimensions
were selected to achieve just this result. In the past I had compared this Geiger counter
against another "bomb shelter" type and obtained surprisingly close readings -
maybe within 10%. The radiation disk gives 1500 CPM when held directly against the Geiger
tubes mylar window and 500 CPM when the lid of a mint can is placed in between (to
simulate the ion chamber walls). The background radiation measured about 13 CPM. Now here
is where the calibration gets a bit "iffy". The disk is large compared to the
Geiger tube window but it is small compared to the diameter of the ion chamber. To make a
long story short, the ion chamber will read low by some factor - maybe 4. What I think it
all means is that the ion chamber gives about 6 mV for a radiation level that causes about
10 CPM in the Heathkit unit corresponding to 0.01 mR/hr. The background radiation should
give a reading just above 2 mV which is about how much the readings wander from minute to
minute. This calibration may be within one order of magnitude!
But now I am hooked. I want an ion chamber that can easily see the
background radiation. To make a bigger chamber, I chose a 4.5" by 4" diameter peanut
can (see below). I also decided to move the FETs into their own compartment. For this
compartment, I used a steel wheel from the center of an electronic component reel. Any
small can would work here, but this piece fits nicely on the bottom of the peanut can and
it had a hole perfect for the 8-pin header that I happen to have in large quantities. A
hole was drilled in the bottom of the peanut can for the electrode. No insulator was used
- just the air gap. You can see the wire sticking up from the FETs in the first picture
and the wire is visible passing through a hole into the chamber in the second picture. The
end of the peanut can was sealed with aluminum foil to keep out air currents and electric
fields but to allow less energetic or larger particles in.




The voltage on the outer can was increased to 22.5 volts by using an old 'B' battery on
the theory that a larger chamber would need a larger field to sweep out the ions quickly
enough.
After power was applied and sufficient settling time went by (about 15 minutes), the
meter reading was seen to be significantly more jumpy. The FETs were the same ones from
the previous chamber and great care was taken to keep everything clean so I immediately
suspected that I was seeing individual ion trails. When I slipped the radiation disk under
the aluminum window, the reading climbed to a whopping 1.5 volts! And what a coincidence!
The Heathkit gives a count of 1500 CPM for this same source when covered with aluminum
foil. (Actually, the foil hardly attenuates the radiation coming from the disk.) So now I
have a direct readout of mR/hr: 1 volt = 1 mR/hr. Unfortunately, I have not yet
corrected for the much larger detector area of the new chamber but it works out to be
nearly 10! So the sensitivity of the new chamber is 1 volt per 0.1 mR/hr which is pretty
sensitive! The radioactive element from a smoke detector was held up to the Geiger counter
and the count soared to about 22,000 CPM but placing a piece of aluminum foil in between
dropped the count to 200 CPM. The ion chamber gave a reading of 200 mV which is in perfect
agreement. But I didn't expect agreement since this source is small relative to the Geiger
tube, also. The mylar window on the Geiger tube blocks the alpha particles some and this
may account for the agreement. These calibrations are really coarse! By using the Geiger
counter to measure the background radiation it was determined that the ion chamber should
be indicating 13 mV but since the zero setting is arbitrary, it was hard to confirm this
level. Reversing the polarity on the outer can caused a shift of about 30 mV (after
several minutes of settling) which is about what is expected if the background is near 15
CPM (plus 15 to minus 15 is a total of 30). The experimental setup is shown below:
More Experiments:
I tried a long chamber made from a section of air duct with a thin wire
stretched between two Styrofoam plugs:
The internal wire was soldered to both ends, the plugs taped in place and the tube stretched to tighten the wire |
.

But then:
Another chamber was constructed with a large cookie tin similar to the
peanut can design above. The performance of this much larger chamber was excellent. A
single Coleman lantern mantle nearly "pegged" the output. The background radiation
gives about 4 mV (400 mV after amplification) which corresponds to 40 fA current.
(Some
CMOS opamps have input current below 40 fA like the LMC6001 and would work fine
without the JFET.) Even though the circuit was given a
low frequency response to reduce 60 Hz response, the meter jitters in response to
individual ion trails. (The superior shielding of the cookie tins would probably allow for
a faster response, if desired, but watch out for circuit instabilities.)
| This tin measures about 10.5" across and about 6" tall (a "regular" height tin should work as well). The center portion of the lid was cut out with scissors to make a frame to hold the aluminum foil window. The circuit is housed in a smaller cookie tin tack-soldered to the bottom. Connections are made via a 5-pin audio connector. |


The electrode looks like a
lasso. |
The electrode is a 5" dia. wire ring mounted to a Teflon standoff and a short piece of stiff telephone wire connects the electrode to the circuitry. The wire passes through a large hole to reduce the chances for leakage currents. The circuitry is a modified version of the first schematic featuring a resistor for the feedback and an op-amp for boosting the output signal. The transistor circuit was also modified to increase the loop gain and improve the stability (see ckt. desc. below). |
| A word of caution: the metal sure looks like ground and a person (um, like me) might start soldering components that go to ground to it. The can will actually be connected to +45 volts or more so the "ground" connections are made above the metal. The only components that connect to the can in the photo are a large yellow cap and a couple of white caps used to support my elevated ground buss. |


The new circuit includes several improvements. The feedback FET is replaced with a
Victoreen 100,000 megohm resistor which is the long glass tube in the photo. A zener diode
was added to the emitter of the 2N4401 to increase the loop gain and a .01 uF Miller
capacitor was added to reduce the amplifier frequency response (for stability and to
reduce 60 Hz gain). An op-amp (OP-07) was added to boost the output by a factor of 100.
The "zero" pot is used to set the output to a few volts since the OP-07 cannot
swing below 1 or 2 volts out without a negative supply. This pot must be able to be
adjusted to the gate voltage and with some FETs the voltage may not go low enough. The
symptom will be a high op-amp output voltage. If so, just lower the 10k resistor or
add a 1k above the pot. An additional zero pot for the meter could be added as in the
first schematic to get a near-zero reading for the background radiation, if desired.
Notice that the drain resistor was reduced to 125k. This value was experimentally
determined by finding the drain current that gives the 2N4117A a near-zero temperature
coefficient. The test circuit is simple: connect a sensitive current meter from +10 volts
to the drain, ground the gate, and connect the source to ground through a 500k pot. The
current is observed at room temperature then the FET is warmed and the current change is
noted. The pot is adjusted until little or no change occurs. I heated the FET by touching
a warm PTC to the can - probably reaching about 65 degrees Celsius and the final current
change was below the current change caused by a 100 uV gate voltage change. (Corresponds
to less than 1 fA ion chamber current for 40 degrees.) Room temperature may vary by +-4
degrees which would correspond to a wander of 0.1 fA which is well below the 40 fA
background current from the chamber. The bias current that gave this wonderful temp-co was
40uA and since the drain resistor will have 5 volts across it, the desired resistor value
is 5/40 uA = 125k. "Your results may vary." Actually, the FETs are surprisingly
stable at all currents and the whole procedure may be unnecessary; just use 125k!
Also consider the circuit used in the CDV-715
mod below. The mod is hard but the circuit is easy. Also, my circuits use
ultra-low leakage JFETs because I have about a thousand of them but there are
also op-amps that can do the job directly. Investigate the LMC6001 (25 fA,
tested!). Just leave the FET and source resistor out of the CDV-715
circuit below and connect the negative input of the op-amp directly to the
sense wire. Don't connect the 33k battery test resistor.
A 22.5 volt battery was insufficient to capture all of the ions but two batteries (45
volts) seemed to do the job - in other words, higher voltage did not result in a higher
reading. Higher voltages may be desired for observing individual events, however, since
the ions will be swept to the electrode faster.
Parts Notes:
| The 4.7uF capacitor should be a non-polar film type with a voltage rating above the voltage used (45 volts in the schematic). A non-polar type allows the voltage polarity to be reversed for experimental purposes. | |
| The 100,000 megohm resistor is a specialty device which may be hard to obtain. | |
| The 2N4117A is an unusual electrometer-grade JFET which has few substitutes. | |
| The other components are not particularly critical. |
Ion Chamber Bias Supply (Battery Topper)
After discovering that one of my bias batteries was jumping
around a few volts, wreaking havoc with the readings, I decided to build a
floating, regulated high voltage supply. The result is a micro-power 110 volt
supply that runs on an ordinary 6V lantern battery. The circuit is similar to my
Geiger counter supply but without the additional voltage multiplier on the
output. Current consumption in a typical ion chamber setup is only 150 uA
(average) so a lantern battery will last a decade, if the shelf life doesn't get
it first. No power switch is needed.

The circuit
has an output filter consisting of a 100k resistor and a .22 uF
capacitor and a typical experiment will have additional capacitance
across the chamber, too. To get rid of the last bit of wander as the
circuit pulses, increase the 100k to 10 megohm since ion chambers don't
draw significant current. I left the value lower in case this device is
used to drive a heavier load and added an external 10 meghom in series
with the output with a 10 uF polyester capacitor to ground right at the
ion chamber. The load can't be too heavy, however. This circuit can
just barely drive a 22 megohm resistor with the full 110 volts and a 10
megohm multimeter will load it down to about 85 volts. The .22 uF will
still bite you when it's charged and a charged 10 uF will really get
your attention! So be careful!
The circuit bolts right onto the terminals of the battery so I
have dubbed this type of circuit a "Battery Topper". Sorry, I couldn't help it!
I actually like the idea of mounting handy circuits right on a lantern battery
for quick lab circuits. Solder lugs were added to make contact with the battery
terminals and narrow nuts were screwed onto the battery first to raise the
circuit up a bit so that the hand wiring underneath doesn't press against the
top of the battery. The circuit was built on a piece of countertop laminate.
Experimenter's Ionization Chamber
Here is a truly simple experimenter's
chamber made from an ordinary cookie tin:


This experimenter's chamber is made from a 4" (10 cm) diameter, 5.5" (14 cm)
tall tin with a tight-fitting lid. The inside of the can is conductive and does
not appear to have the typical plastic coating. A 5-way binding post is mounted
in the center of the can and a 4" (10 cm) wire is suspended from the post inside
the can. The wire length is short enough to insure that it doesn't touch the
lid. Another all-metal binding post and pin are installed in the bottom of the
can, and a sheet of gray insulating plastic is glued into place to keep hastily
constructed experiments from contacting the can. The electrometer circuitry will
be extremely sensitive to stray electric fields, so a shield is mandatory.
Another can previously containing mints is pressed into service:

A pin jack is soldered to the tin so that it can be plugged onto the pin on
the chamber, and tape is applied to the lip of the can so that the pin is the
only connection point. The inside surfaces of the shield are also insulated with
tape to prevent accidental contact with the circuitry. The method of connecting
the shield isn't critical, and a clip lead will also work; the main culprits are
the ever-present, low frequency, line-related electric field and changing
electrostatic fields due to movement near the chamber. The wires from the test
circuitry can simply slip between the shield and the chamber, or a small notch
may be made in the shield to make a little room for a few conductors. The
opening of the chamber may be covered by the original lid, aluminum foil, or
wire screen, depending on the experiment. Leaving the end open will let in too
much stray electric field in most environments.
Here's a simple starter experiment:
The can is connected to the positive
battery voltage through a 4.7k resistor, and the meter is connected between the
collector of the transistor and the positive terminal of the battery. The meter
is on the 1 volt scale for most measurements.

The
transistor is an ordinary NPN Darlington type like the MPSAW45A. The resistor
can be any value above 1k; it simply limits current in the event of a short
circuit. A little piece of double-sided foam sticky tape holds the battery in
position.
When a ray passes through the chamber, several atoms are ionized and the
positive voltage on the can attracts the electrons. The positively charged atoms
wander to the more negative center wire and, upon contact, reclaim their missing
electrons. This process results in a current flow in the base of the transistor
which is amplified by a factor near 30,000. This higher current flows through
the 10 megohm resistance of the meter, producing the indicated voltage. As a
point of reference, a reading of 10 mV would correspond to roughly 200,000
electrons per second, so even weak radioactive sources produce large numbers of
ions.
To observe the background and leakage level, the lid is placed on the bottom,
the top shield is added and the reading is allowed several minutes to stabilize.
The meter settles to a little over 30 mV and exhibits an occasional jump. A
camping lantern mantle known to contain radioactive thorium is place in the lid
of the chamber, and the lid is secured on the open end of the can such that the
mantle is inside the chamber. The meter reading climbs to over 600 mV:

Placing the item to be tested inside the chamber in this manner gives the
ultimate sensitivity, but care must be taken to avoid touching the center wire.
This very simple detector demonstrates how easily an effective radiation
sensor may be made with a minimum of effort. Below is shown another way to build
this simple circuit with even less effort. First, solder the 4" wire directly
onto the base of the transistor:
Drill a hole in the can right in the center of the bottom and epoxy the
transistor, face-down, such that the wire protrudes into the can without
touching the sides. Make sure that no epoxy touches the center lead of the
transistor (base lead). The epoxy is too conductive!
Connect two wires to the collector and emitter leads. The picture shows a
length of solid copper telephone cable used for the connections. The two blue conductors
are connected to the transistor legs and the two orange wires are connected to
the can. The blue wires are given a little slack so that the cable pulls on the
can connection and not on the transistor legs.
The resistor is connected on the other end of the orange wires. A mint tin is
tacked to the coffee can in a couple of places to act as a shield and a wiring
reminder is written on the back of the mint tin. A little optional notch is seen
near the bottom of the mint tin that allows the phone cable to exit the tin
without being pinched. The opening of the coffee can is covered with ordinary
kitchen aluminum foil held in place by an elastic band.


The electrode looks like a
lasso.